understanding active electrodes and impedance
Posted: Wed Apr 14, 2010 3:26 pm
Every time I show someone how to use our Biosemi system they want to know how to measure impedance and seem very worried that they can't. This is especially true of people who have experience with EEG. I'm always a bit unsure about how to answer their questions. Below I've written my latest attempt at explaining why with passive electrodes checking impedance is especially important but with Biosemi active electrodes it is less of an issue (at least that is what I'm assuming). Can you please give me some feedback and more information on this issue? This will be very useful for anyone using the system who may face the impedance issue in dealing with colleagues or reviewers of manuscripts. I apologize if the reasons for this are clearly laid out elsewhere in the technical publications. Unfortunately, I don't have the expertise to really understand those (I have tried) so I'm looking for something on a level (basic college electricity and magnetism. I was a biology major!) that I and other users of the system can grasp more intuitively (but still accurately!).
(1) PASSIVE ELECTRODES - Current flows into/out of the head across the skin/gel/electrode interface and through the electrode wire to/from the amplifier. At various points along this path the current will face some resistance (something that restricts the flow of current). For alternating current (as we have in EEG signals) this resistance is called impedance because in addition to resistance, capaciatance and inductance also affect the ability of AC current to flow. Impedance/resistance is related to voltage and current by Ohm's law; Voltage = Current x Resistance. If the impedance across the skin/gel/electrode interface is high, then this will mean that it will be harder for current to flow to/from the amplifier (especially along a long wire that has lots of chances for interference) and a higher voltage will be necessary in order to get a signal through the wire to the amplifier. Thus, higher impedances result in poorer signal quality because amplitudes will be attenuated. Impedance across the skin/gel/electrode interface is usually lowered by removing the layer of dead skin cells on the surface of the skin.
(2) BIOSEMI ACTIVE ELECTRODES - A "pre-amplifier" is inserted just after the electrode and before the long wire that goes to the amplifier. This amplifier does not give the signal any gain. I'm not sure exactly what happens here? Is common mode (CMS) rejection done here and the difference then transmitted down the wire? Why does the impedance across the gel/skin/electrode interface no longer matter? Why doesn't the same situation as for passive electrodes come up (higher impedance > lower current down the wire > lower measured voltage?) or have I gotten that wrong?
Impedance seems to have a big effect on signal quality in passive systems and thus it needs to be equivalent across channels. Is there some other factor that has a major impact on signal quality in the Biosemi system and does it need to be equivalent across channels. I have noticed that some channels do need to have more gel or a little scratch or two to make the signal APPEAR QUALITATIVELY better (i.e. less noisy, more stable). Is this changing the impedance? It would be nice to have some more QUANTITATIVE measure of what I have done.
I very much appreciate any help that you can give on these issues. I want to understand better how the system is working...
(1) PASSIVE ELECTRODES - Current flows into/out of the head across the skin/gel/electrode interface and through the electrode wire to/from the amplifier. At various points along this path the current will face some resistance (something that restricts the flow of current). For alternating current (as we have in EEG signals) this resistance is called impedance because in addition to resistance, capaciatance and inductance also affect the ability of AC current to flow. Impedance/resistance is related to voltage and current by Ohm's law; Voltage = Current x Resistance. If the impedance across the skin/gel/electrode interface is high, then this will mean that it will be harder for current to flow to/from the amplifier (especially along a long wire that has lots of chances for interference) and a higher voltage will be necessary in order to get a signal through the wire to the amplifier. Thus, higher impedances result in poorer signal quality because amplitudes will be attenuated. Impedance across the skin/gel/electrode interface is usually lowered by removing the layer of dead skin cells on the surface of the skin.
(2) BIOSEMI ACTIVE ELECTRODES - A "pre-amplifier" is inserted just after the electrode and before the long wire that goes to the amplifier. This amplifier does not give the signal any gain. I'm not sure exactly what happens here? Is common mode (CMS) rejection done here and the difference then transmitted down the wire? Why does the impedance across the gel/skin/electrode interface no longer matter? Why doesn't the same situation as for passive electrodes come up (higher impedance > lower current down the wire > lower measured voltage?) or have I gotten that wrong?
Impedance seems to have a big effect on signal quality in passive systems and thus it needs to be equivalent across channels. Is there some other factor that has a major impact on signal quality in the Biosemi system and does it need to be equivalent across channels. I have noticed that some channels do need to have more gel or a little scratch or two to make the signal APPEAR QUALITATIVELY better (i.e. less noisy, more stable). Is this changing the impedance? It would be nice to have some more QUANTITATIVE measure of what I have done.
I very much appreciate any help that you can give on these issues. I want to understand better how the system is working...