Elec. ref. for one biosemi system multiple participants

Post Reply
Chanel
Posts: 17
Joined: Thu Dec 08, 2005 11:03 am

Elec. ref. for one biosemi system multiple participants

Post by Chanel »

Hello,

we plan to record online two participant with one biosemi system. One paticipant using EMG recordings and the other using EEG recording. The main question we have is the question of references.

Is it possible to use both references (EEG CMS&DRL and front panel CMS&DRL) at the same time for different participant ?

Otherwise is there a way to have a common reference for both participants ?

Thanks in advance.

Coen
Site Admin
Posts: 1125
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

It is not possible to measure two subjects simultaneously with only one AD-box. However, the ActiveTwo system offers the unique option to measure up to four subjects simultaneously with daisy-changed AD-boxes (each subject has its own AD-box). Information is found on http://www.biosemi.com/new.htm, scroll down to 26 September 2005.

Best regards, Coen (BioSemi)