Hello,
I would like to record EOG signal in order to analyze the position of the eyes. In the litterature, they recommend a Direct Current Amplifier (because the Alternative Current genetares a drift to zero after each eye movement). Is there a possibility to record the signal in a Direct Current with BioSemi? If yes, how?
Thanks in advance,
Anaïs
EOG - Direct current (DC) versus Alternatice current (AC)
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: EOG - Direct current (DC) versus Alternatice current (AC)
The BioSemi ActiveTwo system has a DC amplifier, signals saved to file are always DC.
Switch the high-pass filter (left sidebar) off to display the signals in DC mode during recording. Note: the filters (as well as all other controls) in the left sidebar do not affect the data saved to file. Bandwidth saved to file is always DC to 1/5th of the sample rate (as indicated in the right sidebar).
Best regards, Coen (BioSemi)
Switch the high-pass filter (left sidebar) off to display the signals in DC mode during recording. Note: the filters (as well as all other controls) in the left sidebar do not affect the data saved to file. Bandwidth saved to file is always DC to 1/5th of the sample rate (as indicated in the right sidebar).