Hi, I am new to this system, and have a question about the trigger inputs.
The trigger box sets the input pins to 3.3V when nothing is connected.
As far as I can tell, the only way to change this is by shorting the inout so that the voltage across the pins goes to 0.
I would like to trigger off of the signal from an accelerometer, or really just any signal, is it possible to do this?
Thank You
Triggering with BioSemi Active 2
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Triggering with BioSemi Active 2
The trigger inputs have an internal 10 kOhm pull up resistor. Your trigger signal must be TTL level (Low = 0V, High is 3.3-5V), and must be able to drive the 10 kOhm resistor (be able to pull the input below approx. 1V).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)