Did you use the unreferenced (raw) signals as basis of your analysis ? The correlation is a measure for the amount of Common Mode in the signals. The CMS/DRL loop becomes less effective at higher frequencies (the loop amplification decreases proportionally to the frequency). So, it is to be expected...
by Coen
Tue May 26, 2009 5:10 pm
Forum: BioSemi Software (LabVIEW programming)
Topic: regarding USB_WRITE function in labview_dll.dll
Replies: 6
Views: 23221

The program reads (variable) parts of the ringbuffer. The data remains available in the ringbuffer until overwritten by new data (when the "clock" has made a full round). The convertxxx vi's are used to adjust the program to the version (Mk1 or Mk2) and sample rate (speedmode) of the connected AD-bo...
by Coen
Fri May 22, 2009 2:48 pm
Forum: BioSemi Software (LabVIEW programming)
Topic: regarding USB_WRITE function in labview_dll.dll
Replies: 6
Views: 23221

Act_Pointer.vi determines whether new data is available in the ringbuffer. If not enough new data is available, the program waits. If enough new data is available, the program reads a new block of data from the ringbuffer.

Best regards, Coen (BioSemi)
by Coen
Fri May 22, 2009 1:45 pm
Forum: BioSemi Software (LabVIEW programming)
Topic: regarding USB_WRITE function in labview_dll.dll
Replies: 6
Views: 23221

The USB_WRITE call is used for two functions: - start and stop acquisition, 1st byte of the 64 byte array is used, in LabVIEW, I used the values 1 and 0 to start/stop the acquisition (actually, only the LSB of the byte is read by the USB receiver to start and stop) - sent output triggers to the trig...
by Coen
Thu May 21, 2009 3:58 pm
Forum: EEG data analysis
Topic: Error 6004: File Size in Header
Replies: 1
Views: 6083

The usual cause for a wrong file size in the header is that the recording was not stopped normally (but instead by a PC crash, full harddisk, etc.). If the file is basically correct, and the problem is just in the header, then you can repair the file with the tool available from http://www.biosemi.c...
by Coen
Tue May 19, 2009 1:23 pm
Forum: BioSemi Software (ActiView executable)
Topic: error 5000 remains
Replies: 4
Views: 8385

If the optical link light remains off, then I suggest to replace the optical fiber.

The flashing USB light indicates that the communication between PC and receiver works.

Best regards, Coen (BioSemi)
by Coen
Tue May 19, 2009 12:42 pm
Forum: BioSemi Software (ActiView executable)
Topic: error 5000 remains
Replies: 4
Views: 8385

f nothing has been changed to the PC and AD-box (speedmode ?), then the USB cable and optical cable are the prime suspects. - Is the "optical link" light on the USB receiver on to indicate that data is received ? - What happens with the "USB link" light on the receiver when acquisition is started ? ...
by Coen
Mon May 18, 2009 3:00 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

As far as I know, the basic ActiView code can be used. However, the problem with LabVIEW RT is that it needs to run on dedicated real time (National Instruments) hardware. An interface between the ActiveTwo and this special real time hardware is not available at the moment. An option (using our stan...
by Coen
Mon May 18, 2009 12:55 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

A new version of the zip file (same name, now 3.7 Mb) is available on the same link as above, the full source code is now added to the zip. Follow the "Triggers Boolean" output array from subvi "TrigProcess.vi", and you'll find an added USB_WRITE function call (lower right section of the main diagra...
by Coen
Thu May 14, 2009 3:56 pm
Forum: BioSemi Hardware
Topic: Measuring with 2 different electrode sets
Replies: 2
Views: 6262

Old electrodes may produce more (low frequency) noise than newer ones. Electrode wear cannot significantly influence signal amplitude. You can easily check this with a "two bucket test", as described in the guidelines for use of the ActiveTwo system. In other words: you may encounter differences in ...
by Coen
Wed May 13, 2009 2:19 pm
Forum: BioSemi Software (ActiView executable)
Topic: heartrate measure
Replies: 1
Views: 5196

Save only the relevant channels to file (do not save "flat" channels), and reduce the sample rate for just enough analog bandwidth (but not more).

Best regards, Coen (BioSemi)
by Coen
Tue May 12, 2009 3:25 pm
Forum: BioSemi Hardware
Topic: Re-referencing and MMN/N400
Replies: 3
Views: 7518

I don't know an "official" coordinate. However, the coordinates of the nose can be quite easily extrapolated from the existing ones. The azimuth is surely 90 degrees (same as Fpz, Fz, etc), the inclination should be something around 138 degrees (Fpz is on 92, arc length from nose to Fpz is approx eq...
by Coen
Mon May 04, 2009 1:24 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

A special version of ActiView is ready for download from http://www.biosemi.com/download/ActiView606-Lores.zip . This version routes the response signals from input trigger lines 9 and 10 to trigger output lines 1 and 2 (pin 17 and 18, see http://www.biosemi.com/faq/trigger_signals.htm). Please note...
by Coen
Mon Apr 27, 2009 7:59 pm
Forum: EEG data analysis
Topic: Suspicious about the electrode locations file
Replies: 1
Views: 5888

Why is the value not possible ? Why should the values be 0,10 ? If you claim that there is an error, than please provide a justification for your claim.

The polar coordinates are given in degrees, and used successfully by hundreds of our customers.

Coen (BioSemi)
by Coen
Mon Apr 20, 2009 4:03 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

The answer is above, ActiView should be reprogrammed to sent out triggers, and a special cable should be constructed.

Best regards, Coen
by Coen
Mon Apr 20, 2009 3:09 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

The provided response box connect to the AD-box. The triggers are directly available in ActiView. Make sure you have set the CFG to see the triggers. The trigger cable is not involved.

I was referring to sending triggers from the acquisition PC to output trigger lines.

Best regards, Coen (BioSemi)
by Coen
Mon Apr 20, 2009 2:19 pm
Forum: BioSemi all other topics
Topic: Standard BioSemi PArallel Port Cable Incompatibility
Replies: 12
Views: 16437

The LabVIEW code of ActiView Light (http://www.biosemi.com/download/ActiView604-Light.zip) shows an example of how to sent codes to trigger port (15 lines available). You will have to make the appropriate cable yourself according to the information on http://www.biosemi.com/faq/trigger_signals.htm B...
by Coen
Tue Apr 07, 2009 1:32 pm
Forum: BioSemi Hardware
Topic: Strange looking drifts through all channels
Replies: 5
Views: 9775

When water or gel enters the electrode, they will stop working, and they won't work again (blue light keeps blinking). Problems with leaking electrode have never been reported to us. A possible cause is pollution of the Ag-AgCl tips with traces of non-silver metal. However, the low frequency noise w...
by Coen
Tue Apr 07, 2009 11:45 am
Forum: BioSemi Hardware
Topic: Strange looking drifts through all channels
Replies: 5
Views: 9775

The artefacts in your file ("bump" on most channels from t=76 to t=80) look like slow waves due to sweating, as discussed in thread viewtopic.php?t=28.

Best regards, Coen (BioSemi)
by Coen
Mon Apr 06, 2009 12:29 pm
Forum: BioSemi Software (ActiView executable)
Topic: TTL Pulse width recognition
Replies: 3
Views: 8686

The trigger lengths values refers to the trigger signals coupled to the trigger port on the USB receiver.

When triggers are controlled with the function keys, the line remains high approximately as long as the key is pressed.

Best regards, Coen (BioSemi)
by Coen
Wed Mar 25, 2009 3:09 pm
Forum: BioSemi Hardware
Topic: recording only external (eye) electrodes
Replies: 3
Views: 6816

If you could record previously with many headcap channels, then there should be no problem at all with recording just the 8 EXG electrodes. You will just need the separate CMS/DRL lead (connection to the front panel of the AD-box).

Best regards, Coen (BioSemi)
by Coen
Wed Mar 25, 2009 3:03 pm
Forum: BioSemi Hardware
Topic: Re-referencing and MMN/N400
Replies: 3
Views: 7518

In general, you can check your referencing step because the reference trace should be flat after referencing (everything zero, no noise). In your example, the EXG5 trace should be flat after referencing all channels to EXG5. The CMS/DRL loop cancels out after referencing In general, the reference ch...
by Coen
Fri Mar 13, 2009 5:25 pm
Forum: BioSemi Hardware
Topic: recording only external (eye) electrodes
Replies: 3
Views: 6816

A ringbuffer overflow means that the PC is not able to keep up with the incoming data stream. Basically, the problem can only be solved with a faster processor in your acquisition PC A CMS and DRL electrodes is always necessary, also if you only use the flat electrodes (EX inputs). The ringbuffer ov...
by Coen
Fri Mar 06, 2009 11:16 am
Forum: BioSemi Hardware
Topic: Diagnose a Broken Electrode (with pictures!)?
Replies: 4
Views: 10601

The signal trace suggest a bad contact between electrode tip and skin (not enough gel in the holder). The electrode itself is most probably OK. A defect electrode (broken wire) will trigger the safety circuits, and signal on all electrodes will be lost (blue light flickering). Another possibility is...
by Coen
Wed Mar 04, 2009 12:30 pm
Forum: BioSemi Hardware
Topic: usb reciever output
Replies: 1
Views: 5230

With the Jazz enabled (output trigger is 16 High), the special Jazz version of the USB receiver adds 4 extra channels to the data stream. Inspect the LabVIEW code of the subvi named Act_Jazz.vi to learn about how the Jazz information is embedded in the 4 extra channels (it's quite complex). With the...
by Coen
Fri Feb 27, 2009 2:57 pm
Forum: BioSemi Software (ActiView executable)
Topic: Additional Trigger Question
Replies: 6
Views: 12395

The trigger cables as currently delivered by BioSemi already have triggers 9-16 grounded (connected to pin 37). This is not true for our older cables (delivered before 2007), were triggers 9-16 are floating (and pulled high by the pull-ups in the receiver). The change of the 24-bit value in the firs...
by Coen
Thu Feb 19, 2009 5:36 pm
Forum: BioSemi Software (ActiView executable)
Topic: Additional Trigger Question
Replies: 6
Views: 12395

The connection of pin 9-16 to pin 37 is indeed done by soldering a short wire. The pin layout can for example be found on http://pinouts.ru/connector/37_pin_D-SU ... ctor.shtml (pin 1 is also seen inside the connector).

Best regards, Coen (BioSemi)
by Coen
Fri Feb 13, 2009 12:01 pm
Forum: BioSemi Hardware
Topic: optic fiber cable
Replies: 3
Views: 7508

There are no technical problems at all with using a duplex version of the optical fiber. In addition, the most duplex cables are easily split into two separate cables (we have actually done this ourselves during supply problems of the single optical fibers).

Best regards, Coen (BioSemi)
by Coen
Fri Feb 06, 2009 2:38 pm
Forum: BioSemi Software (ActiView executable)
Topic: BDF Record Length
Replies: 2
Views: 6306

ActiView uses a fixed record length of 1 second for the BDF files. However, changing the open-source LabVIEW code yourself to achieve a different record should not present too much problems.

What would be the advantage of a different record length ?

Best regards, Coen (BioSemi)
by Coen
Tue Feb 03, 2009 4:37 pm
Forum: BioSemi Hardware
Topic: AD-box problem
Replies: 2
Views: 5991

Test the system with the electrodes in salt water as described in the manual, and discuss the results with BioSemi directly. An AD-box problem is very unlikely.

Best regards, Coen (BioSemi)