- Thu Sep 23, 2021 7:20 pm
- Forum: BioSemi Hardware
- Topic: CM out of range with more than 64 electrodes
- Replies: 3
- Views: 14628
Re: CM out of range with more than 64 electrodes
Hi Coen, Thanks a lot for your reply. This does indeed appear to be the issue - only channels 1-64 are installed. We have multiple AD-boxes, all of them with only 64 channels installed. However, they do all have the daisy chain input. Is it possible to daisy chain two boxes together in order to reco...
- Wed Sep 22, 2021 6:13 am
- Forum: BioSemi Hardware
- Topic: CM out of range with more than 64 electrodes
- Replies: 3
- Views: 14628
CM out of range with more than 64 electrodes
Hello, For a study in our lab, we've created custom EEG caps with the standard 64-electrode 10-20 system layout plus additional electrodes to oversample a particular region of interest. However, we've found that when adding a third band of electrodes in the C slot on our Biosemi amplifier, the blue ...