Search found 5 matches
- Mon Oct 01, 2018 11:01 am
- Forum: EEG data analysis
- Topic: ABR Peaks Positive or Negative
- Replies: 0
- Views: 36049
ABR Peaks Positive or Negative
Dear Coen, we are measuring ABRs in our lab with BioSemi ABR electrodes. We finally have pretty clean ABR measurements. However, now that we are analysing data from a study, I am doubting whether we are labeling the peak in the right way. The wave V latency is quite early with 90dB clicks if we inte...
- Thu Jun 07, 2018 5:20 pm
- Forum: BioSemi all other topics
- Topic: Unwanted triggers during ABR measurement
- Replies: 7
- Views: 39782
Re: Unwanted triggers during ABR measurement
The problem may be that your trigger generator does not switch the separate lines (bits) within one sample period (62 microsecond). Try using trigger values that switch only one line, e.i. 1,2,4,8,16,32,64 and 128. High (unconnected) lines 9-16 do not cause false triggers. Best regards, Coen (BioSe...
- Tue Apr 24, 2018 5:04 pm
- Forum: BioSemi all other topics
- Topic: Unwanted triggers during ABR measurement
- Replies: 7
- Views: 39782
Re: Unwanted triggers during ABR measurement
Hi Coen,
your suggestion resulted in zero unwanted triggers with 24000 samples. That is great. However, we need more than 8 trigger codes. Do you have any idea how we could make this happen?
Best,
Nathan
your suggestion resulted in zero unwanted triggers with 24000 samples. That is great. However, we need more than 8 trigger codes. Do you have any idea how we could make this happen?
Best,
Nathan
- Fri Apr 13, 2018 9:35 pm
- Forum: BioSemi all other topics
- Topic: Unwanted triggers during ABR measurement
- Replies: 7
- Views: 39782
Re: Unwanted triggers during ABR measurement
Thanks for the quick reply and suggestion. I will check whether this resolves the issue coming week and report back.
Kind regards,
Nathan
Kind regards,
Nathan
- Fri Apr 13, 2018 11:19 am
- Forum: BioSemi all other topics
- Topic: Unwanted triggers during ABR measurement
- Replies: 7
- Views: 39782
Unwanted triggers during ABR measurement
Dear sir, madam, recently we have acquired the ABR cord and input and have been able to record decent ABRs using a stimulus with fixed intensity and a single triggercode (sent synchronously with the click onset). We are now also able to present clicks with different amplitudes and different triggerc...