Search found 3 matches

by fernandapgj
Wed Nov 18, 2015 5:10 pm
Forum: BioSemi Hardware
Topic: trouble with CM in Range
Replies: 10
Views: 24034

Re: trouble with CM in Range

Thank you for your prompt response. I ran the saline water bucket test yesterday and found some faulty electrodes: the two reference electrodes and A27 for branch A, and B3, B17 AND B29 for branch B. Using the accessory CMS/DRL cable and leaving the afore-mentioned electrodes out of the water, the b...
by fernandapgj
Tue Nov 17, 2015 8:08 pm
Forum: BioSemi Hardware
Topic: CM in range
Replies: 0
Views: 11225

CM in range

Hello. My team and I are having the following problem while using our 64 electrode BIOSEMI recording system: As stated on previous messages in this thread, whenever we connect our A branch, the blue light starts blinking. When doing the saline water bucket test, we got it steady for a few seconds bu...
by fernandapgj
Tue Nov 17, 2015 8:07 pm
Forum: BioSemi Hardware
Topic: trouble with CM in Range
Replies: 10
Views: 24034

Re: trouble with CM in Range

Hello. My team and I are having the following problem while using our 64 electrode BIOSEMI recording system: As stated on previous messages in this thread, whenever we connect our A branch, the blue light starts blinking. When doing the saline water bucket test, we got it steady for a few seconds bu...