Search found 5 matches
- Wed Apr 07, 2010 2:56 pm
- Forum: BioSemi Hardware
- Topic: trouble with CM in Range
- Replies: 10
- Views: 35441
Blue light is blinking with all new sets of electrodes
Hello, We bought all new sets of electrodes a few month ago, and they were working fine for a while (one experiment per week or so). But since a few weeks ago, we have got the same problem that blue light starts to keep blinking and we couldn't collect any data... As the electrodes are pretty new, w...
- Fri Feb 19, 2010 12:24 am
- Forum: BioSemi Hardware
- Topic: trouble with CM in Range
- Replies: 10
- Views: 35441
Thank you very much for your very quick help! We checked all the electrodes and found that two of EX electrodes cause the blue light blinking. Now, it makes complete sense because we only bought new A,B,C,D and CMS/DRL electrodes, and kept using old EX1-8 electrodes. We really appreciate for this an...
- Thu Feb 18, 2010 5:54 pm
- Forum: BioSemi Hardware
- Topic: trouble with CM in Range
- Replies: 10
- Views: 35441
Re: trouble with CM in Range
We had exactly the same trouble with the following post, and we bought a new set of electrodes (A,B,C, and D) a few month ago. But yesterday, the same trouble happened. The Blue lights (for CMS/DRL electrodes) on the AD-box was blinking and we couldn't collect EEG data. So we tried a salt-water buck...
- Tue Oct 06, 2009 5:07 pm
- Forum: BioSemi Hardware
- Topic: All offsets exceed +262 mV and CMS/DRL light blinking
- Replies: 2
- Views: 9991
- Mon Oct 05, 2009 9:31 pm
- Forum: BioSemi Hardware
- Topic: All offsets exceed +262 mV and CMS/DRL light blinking
- Replies: 2
- Views: 9991
All offsets exceed +262 mV and CMS/DRL light blinking
We have been having a problem in our lab with the offsets exceed +262 mV and CMS/DRL light blinking. When the ERP starts, all the electrodes are fine, and measuring in an appropriate range. We first incountered the problem when trying to correct the C1 pin electrode offset. After pulling out C1 and ...